I am validating the USP 467 method, new to our lab.
Using Thermo GC and Headspace equipment along with Chromeleon software.
Our question is in regards to the final sample calculation.
We analyzed a calibration curve, 5-5000 ug/ml in DMF. We placed 1ml of each standard in the headspace vials.
For our sample, a dry crystal, we weighed 50mg into 1ml DMF into the headspace vial.
Our on column concentrations of our standards would be expressed in ug, as for example, the 100ug/ml standard, using 1ml, would be 100ug.
We then are taking the on column concentration of each analyte in the sample and dividing by the exact weight (i.e. 0.053mg).
So for example, an on column result of ethyl acetate of 800 ug/ml (ug) would be divided by 0.053g to give us a results of 15,094 ug/g.
We would like to verify that this calculation is correct. Thank you