Analyte degradation during chromatographic analysis

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I tested a rather structurally complex compound with 2 pPenol rings connected by an etheryl bridge: Phenyl--CH2--Oxygen--CH2--Phenyl and both phenols being chlorinated at one position each though at different carbons on the ring.

Upon chrom analysis I observed two peaks--could one be the break-away structure from the parent compound? When fractured at the --0-- site of the molecule? The Tr are at 8.5 mins and 9.3 mins. Given this veryclose association in Tr's, is this hypothesis tenable? To my knowledge, LC does not degrade compounds in situ.
Jumpshooter

This bond is not terribly sensitive to hydrolysis. Plus, if your parent would fall apart at this location, you should get the parent and two product peaks, the right side and the left side. Plus, commonly, the retention of analytes with similar components but roughly two-fold difference in MW is more different than what you are observing.

I think the answer is no...
2 posts Page 1 of 1

Who is online

In total there are 378 users online :: 1 registered, 0 hidden and 377 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 377 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry