retention order for VOC mixture?

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
Hi,
I need retention order for volatile organic carbon mix for GC-FID analysis. Or can you suggest me some components so that I can identify location easily. I have some mixtures from different companies?

Can you share some chromotogram or retention order or any other suggestion.







Benzene, Bromobenzene, Bromochloromethane, Bromodichloromethane, Bromoform, Bromomethane, sec-Butyl benzene, tert-Butyl benzene, n-Butyl benzene, Carbon disulfide, Carbon tetrachloride, Chlorobenzene, Chloroethane, Chloroform, Chloromethane, 2-Chlorotoluene, 4-Chlorotoluene, Dibromochloromethane, 1,2-Dibromo-3-chloropropane, 1,2-Dibromoethane, Dibromomethane, 1,2-Dichlorobenzene, 1,3-Dichlorobenzene, 1,4-Dichlorobenzene, Dichlorodifluoromethane, 1,1-Dichloroethane, 1,2-Dichloroethane, 1,1-Dichloroethene, trans-1,2-Dichloroethene, cis-1,2-Dichloroethene, 1,2-Dichloropropane, 1,3-Dichloropropane, 2,2-Dichloropropane, 1,1-Dichloropropene, trans-1,3-Dichloropropene, cis-1,3-Dichloropropene, Ethyl benzene, Hexachlorobutadiene, Hexachloroethane, Isopropylbenzene, p-Isopropyl toluene, Methyl tert-butyl ether (MTBE), 4-Methyl-2-pentanone (MIBK), Methylene chloride, Naphthalene, Nitrobenzene, n-Propylbenzene, Styrene, 1,1,1,2-Tetrachloroethane, 1,1,2,2-Tetrachloroethane, Tetrachloroethene, Toluene, 1,2,3-Trichlorobenzene, 1,2,4-Trichlorobenzene, 1,1,1-Trichloroethane, 1,1,2-Trichloroethane, Trichloroethene, Trichlorofluoromethane, 1,2,3-Trichloropropane, 1,2,4-Trimethylbenzene, 1,3,5-Trimethylbenzene, Vinyl chloride, p-Xylene, o-Xylene, m-Xylene
Why are you trying to run a TO-15/524/8260 mix on a FID? Buy a MS. If you're dedicated to being a cheap SOB throw your column and analytes into restek's EZchrom, but you should buy individual standards of each compound for analog detectors to establish RT.
RESTEK usually has chromatograms for the VOC's on different columns. You would need two columns for positive identification of peaks. I agree with the last poster. You need a more specific detector. GC-MS would be the best.
Why are you trying to run a TO-15/524/8260 mix on a FID? Buy a MS. If you're dedicated to being a cheap SOB throw your column and analytes into restek's EZchrom, but you should buy individual standards of each compound for analog detectors to establish RT.
Thanks for the reply. I have just started. Company have ms and fid. however, they have only one column. when column passed ms to fid, I have to cut about 30 cm. Therefore I cant use old calibration on fid, plus I don't trust old calibration retention order since ther choose very far away locations for p and m xylene. and ms works for PAH almost non-stop.

Thanks anyway.
Right... let either the $80,000 GC or the $300,000 GCMS sit for want of a $600 column.

You don't need a new column nearly as much as you need new management (and an additional GCMS).
Thanks,
DR
Image
https://docs.accustandard.com/docshare/ ... R2-10X.pdf This chromatogram might be helpful, but I'd also highly recommend using GC-MS instead of GC-FID.
6 posts Page 1 of 1

Who is online

In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: evie and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry