Hollow » Tue Jul 16, 2024 9:12 am
Hi
Just calculate the volume pumped (Flowrate*time)+ and divide it by the colum-volume (r^2*pi*L*)++
Keep that ratio the same for new column dimensions.
Then calculate the time needed for a certain flowrate.
+: if you ramp up the flow, take this into account: draw a chart flowrate vs. time and calculate the area under the curve
++: the true column volume will be a fraction of that (inter particle volume) but for a short approximation and scaling the geometrical volume will do the job.
Are the 180 min experimentally determined or just general procedure?
I would ramp the flow relativley fast to the needed flow rate then equlibrate with this flowrate until the needed volume has been pumped.
You may also test how much equilibration is needed by making injections of a test mix in the same solvent and see if retention time doesn't change between two consecutive injections.