vmu » Fri Jun 07, 2024 4:06 pm
The peak area with UV absorbance detection depends on:
- the mass of the analyte injected to the column (i.e., the analyte concentration in the sample multiplied by the injection volume); the higher the mass, the larger the peak area unless the detector is overloaded or unless the mass is below the injected mass detection limit;
- the flow cell optical length;
- the wavelength (both its nominal value and the spectral bandwidth);
- the units (AU*min, mAU*min, mAU*s, V*min, mV*min, mV*s, "counts"*min, ...; here the x axis of the chromatogram is in time units);
- the flow rate (if the x axis of the chromatogram is in time units); since the time needed for the analyte zone to pass the detector cell is the reciprocal of the flow rate, the peak area is strictly inversely proportional to the flow rate as well; if the x axis of the chromatogram is in eluent volume units, the peak area (e.g., in mAU*ml) is independent of the flow rate.
The peak area with UV absorbance detection is independent of:
- the analyte band broadening (if the peak is high enough in comparison with the signal detection limit);
- the lamp intensity since the absorbance is the log of the reciprocal of transmittance (log(I0/I)).
The peak height with UV absorbance detection depends on:
- the mass of the analyte injected to the column;
- the flow cell optical length;
- the wavelength;
- the units (AU, mAU, V, mV, "counts");
- the analyte band broadening (and all the factors affecting it, namely the column, the extra-column path in the instrument, the mobile phase, the retention factor, the temperature, the flow rate,...); in contrast to its effect on the peak area, the flow rate influences the peak height only via its effect on the band broadening (see the van Deemter equation).
The peak height with UV absorbance detection is independent of the lamp intensity for the same reason as the peak area.