Seeking your sage advice on a noise issue with our Agilent 6890N. Last time we used it 4 months ago for cis-trans analysis with a SP-2560 column it was working beautifully. We left it on in a low-temp sleep (all thermal zones 30-40 C, low column flow and make up gas). This approach has served us well over many years.
Bringing it up, we're confronted with high signal noise at low temp. with signal decreasing with increasing temperature; that is we see extremely noisy peaks for compounds early, which completely disappear. Background level is typical. Here's what we've tried:
Gases: Changed air from Zero Air generator to tank; N2 and H2 carrier and fuel are from tank and generator, respectively. Those gases are shared with other GCs and they're all performing as usual.
Detector: Swapped jet and entirely collector assembly, which included new teflon collector isolators. Inspected electrometer contact spring, tightened all connections. Nothing obviously wrong.
Inlet: Replaced septum, O-ring, liner. Nothing out of the ordinary. Inlet is very clean.
Leaks: Checked everywhere for leaks - column, all fittings everywhere, etc.
Any ideas? Faulty electrometer? I'm hopefully linking an example of the problem. Green is previous, normal. Ignore the cycling baseline drift. That's from the Zero Air generator cycling. Blue is current problem. This is of the same sample mixture.
Here's the link to the hosted chromatogram pic:
https://imgur.com/a/ykc3svm (couldn't get the image to directly display)