- Posts: 362
- Joined: Mon Jul 30, 2018 11:58 pm
For all intents and purposes it's running great and consistently passes BFB tune with rarely needing more intervention than just running "BFB Target Tune." The only issue I sometimes have is BFB target getting 176 to 104% or so(relative to 174), and even if initial tune passes it will go there after ~48 hours, but that's yet to be a problem that's unsolvable. I'll mention too that I'm currently running helium. Column is a CP Select 624 .250mmx30m, and I'm running at 1.2mL/min. Hi-vac is reliably around 1.3x10^-5 torr(gauge) at analytical flow rates.
For as long as I've had this instrument in service, even when all my other numbers looked good, N2/H20 has always been high. Right now it's running ~200%. I've done everything I can think to rule out a leak at the MS itself including a fresh TINY amount of Alpezion-L on the door and vent valve O-rings(and I've also replaced the vent valve O-ring). I've used dust-off to check every potential leak source at the MS, and I can't seem to find it. BTW, I should mention that this problem has carried across a change from a diff pump to a turbo pump(and accompanying vac manifold change).
I'm chasing the last little bit of sensitivity, and am afraid that my high N2 levels are holding me back.
I'm paying a bit of attention to the inlet now, and am wondering if I should start chasing issues there. I did try today snugging up the column nut. I was watching absolute counts of N2 on manual tune/scan and did see it drop after doing that, but tried several other things including playing with the tightness of the septum nut(new septum, never pierced for this) and didn't see a ton of difference. The gold seal, liner, and liner O-ring are all relatively new.
One thing I've never really paid attention to is the split line, which I know is sort of a consumable. I know Agilent sells these pre-cut, but is there any reason why I can't make my own? I have plenty of pre-washed 1/8" copper tubing on-hand, and it looks like nothing but a cut piece of tubing with Swagelok fittings on the end.
Aside from this, is there anywhere else I should be looking for a potential leak?