HILIC Acquity BEH amide column and NH4F
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
I have found various papers describing ammonium fluoride as a promising additive to the mobile phase in HILIC chromatography. However, I have not been able to find anyone reporting using it on an Acquity BEH amide column. It is also not listed in the column manual as one of the possible additives. I have concerns about its suitability for this column, is it compatible?
-
- Posts: 140
- Joined: Thu Jan 07, 2021 6:45 pm
Hi helenaplesnik,
To preface, it should be completely fine to use this buffer in your application. But, I have found an article about using this buffer in HILIC systems here:
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC6950006/
They literally used the exact column that you are using. My one warning would be to pay close attention to the minimums of aqueous solvent that you are using as you may see precipitation of your salt at high-organic percentages. I really hope this helps! If you cant get access to the article, it is open source, let me know and I will send you over a copy!
Have a good one,
Tyler Smith.
To preface, it should be completely fine to use this buffer in your application. But, I have found an article about using this buffer in HILIC systems here:
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC6950006/
They literally used the exact column that you are using. My one warning would be to pay close attention to the minimums of aqueous solvent that you are using as you may see precipitation of your salt at high-organic percentages. I really hope this helps! If you cant get access to the article, it is open source, let me know and I will send you over a copy!
Have a good one,
Tyler Smith.
-
- Posts: 1
- Joined: Fri Jan 26, 2024 7:15 am
Really useful information for me. Thank you very much.Hi helenaplesnik,
To preface, it should be completely fine to use this buffer in your application. But, I have found an article about using this buffer in HILIC systems here:
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC6950006/ doodle cricket
They literally used the exact column that you are using. My one warning would be to pay close attention to the minimums of aqueous solvent that you are using as you may see precipitation of your salt at high-organic percentages. I really hope this helps! If you cant get access to the article, it is open source, let me know and I will send you over a copy!
Have a good one,
Tyler Smith.
3 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 3 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science