Well, this is better. At least we have some solid information on the mobile phase. You also seem to be running a gradient, but you kept this information still secret...
You dissolve the sample in water. This is usually a no-no in normal phase chromatography, since the injection of water will lead to peak distortion and - on the long run - to a deactivation of the column. However, your mobile phase contains some water. With other words, the person who developed the procedure was aware of these potential problems and added water to the mobile phase to develop and maintain an equilibrium with the water. The surface of the silica contains now a controlled amount of water, and good method reproducibility can be achieved with such a method.
Since you have a water-controlled surface, there is nothing wrong with injecting an aqueous sample. However, there can be other issues. There will be a limited solubility of your aqueous sample in the mobile phase, which contains a large proportion of DCM. It is possible that with a sufficiently large injection volume, you will get a plug of sample that migrates through the column without ever seeing the stationary phase. you need to watch out for such an event as you increase the sample volume. Also, pressure spikes are possible.
So, as Hans said: go for it! But also be aware that there may be some issues.