The baseline of the above chromatograms were expanded and examined. They looked fine or consistent.
The lambda max (275 nm) was examined in the absorbance spectrum of Guanine and concluded not likely to be the cause.
So, we are going to re-inject the trouble control sample five times with a known control with supporting data into the Waters HPLC to ensure that the problem was not the sample and something else. We would be able to see the results sometime this week.
If sample is not the cause, then the issue can be the HPLC system. Your suggestion is needed.
Instrument: Waters 717 HPLC with a PDA detector
Mobile phase: 1 % Methanol in 50 mM Sodium Acetate, pH 4.5
Column: Phenomenex C18, 150 x 4.6, 5 micron, 110A
Flow rate: 1 mL/min
RT of peak: 7 min
The HPLC is old, so plumbing can be an issue with a 1% mobile phase on a C18?