Agilent 6130MSD Quad Fault

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello,

I've been spending time getting my university's recently donated LCMS up and running, and I've been obtaining good results. This morning, however, the LCMS has been throwing errors every few minutes that's disrupting runs and ability to auto tune, and interrupts diagnostic tests, some of which are failing.

I get both "Difficulty with ion optics electronics[0200]" and "DC Quad Fault [0209]."

I have experience with opening up 5975/5977 sources and cleaning the optics, but I don't have any LCMSD maintenance experience.

Any ideas where to start?

Thanks!
This sounds like it could be issues with a quad driver board, you may need to find a spare to swap out. Also good to get to the electronics and make sure that no wires were jostled out of place during transport that could cause an intermittent connection fault.

Alternatively you can also try cleaning the ion optics and quads to see if that helps things, but it's fiddly and we usually just get the vendor to do that because it can go quite wrong if you use the wrong solvents/cloths/etc.
2 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry