flow path to uv and mass spect are very similar.
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Assuming that you are using the same mobile phase, column, flow rate, etc. My guess would be that perhaps your source is not being flushed out quickly enough by the exhaust pump? What instrument and source type are you using?
Check the connection tubing between the UV and the MS. If the tubing has the diameter of a pipeline and its length is from Alaska to Texas, the peaks will become a bit wider and have tailing.![]()
We try to set up our systems with minimal length of tubing between the UV and the MS. Also, the volume of the UV cell itself can be a factor.
Then the combination of the connection tubing to the MS together with the split ratio might be the culprit. Shorter and smaller i.d. tubing to the MS might be useful.
Most things that I can think of that would make wide peaks in the MS while the UV peaks are narrow do not create tailing. Since you say that the MS peak is not only wider but also more tailing, I am inclined to think that it is bandspreading in the tubing.
You can try this all out by yourself, and confirm or dismiss my proposition that the tubing is the possible source of the problem. Disconnect the column and inject the same sample under the same mobile phase conditions and flow rate with the same setup, just without the column.