TD-GC/MS, foreline pressure problem

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello, We are using TD-GC/MS in our lab (TD-100, 7890A, 5977A).

When analyzing some sorbent tubes, the foreline value increases from 80 to 300.

And the light on the filament goes out (MS off).

At first, I thought it was due to the high concentration of the sample.

However, it happened with a conditioned sample [there is no problem with the conditioner].

The engineer told me that the diffusion pump is weaker than the turbo pump, so if you change the pump, it shouldn't happen again.

I don't want to replace the pump because it is too expensive.

Is there any other alternative or am I missing something?

Just wondering if anyone else has experienced the same thing I have.
What you describe would indicate that the flow through the analyzer is too high. What is the flow through the TD when desorbing? If you are trying to send more than 3ml/min through the column then you may be overloading the vacuum system.

Can you set a solvent delay until after the vacuum returns to normal? That may be a solution if it is just a sudden jump in pressure as it desorbs.
The past is there to guide us into the future, not to dwell in.
I confirm that with a turbo molecular pump I never had problems.
Good luck
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 3 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Gregorygex, Majestic-12 [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry