I have a Varian HPLC Prostar pump that works for about 2 minutes and then completely ceases to function. No errors appear on the screen. Ideas and advice?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
How long does the pump have to sit before you can restart and have it run for another two minutes?
If you're not seeing any diagnostics, the odds are it's a board-level electronic problem, and as such probably not a user-serivceable item.
If you're not seeing any diagnostics, the odds are it's a board-level electronic problem, and as such probably not a user-serivceable item.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
My experience with 1090 pump of this type of fault is either check valve's malfunction or over pressure.
-
- Posts: 2
- Joined: Tue Jan 12, 2010 10:19 pm
The pump has to sit about a half hour to an hour before a restart will get it working for another 2 minutes.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
I'd still bet on an electronic failure, either on one of the boards or a power supply. Sometimes a marginal component / contact / solder joint will work OK when cold, but fail as the unit warms up.
I don't have any hands-on experience with that particular pump, so all I can make are general suggestions:
Start by checking the simple things first:
- loosen the fitting at the outlet of the outlet check valve, then re-tighten it and see if the pump will start up again immediately. If it does, then you either have a blockage with overpressure, or a pressure transducer that is reading incorrectly (too sensitive).
- go in and disconnect / reconnect every electrical/electronic cable you can find. Hopefully, that will be enough to rub away any corrosion.
If neither of those solve the problem, it's time to get a service technician in to look at it. If you have more time than money, and if you have another system of the same model, you might want to try swapping circuit boards to pin down the source of the problem in more detail.
I don't have any hands-on experience with that particular pump, so all I can make are general suggestions:
Start by checking the simple things first:
- loosen the fitting at the outlet of the outlet check valve, then re-tighten it and see if the pump will start up again immediately. If it does, then you either have a blockage with overpressure, or a pressure transducer that is reading incorrectly (too sensitive).
- go in and disconnect / reconnect every electrical/electronic cable you can find. Hopefully, that will be enough to rub away any corrosion.
If neither of those solve the problem, it's time to get a service technician in to look at it. If you have more time than money, and if you have another system of the same model, you might want to try swapping circuit boards to pin down the source of the problem in more detail.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts
Page 1 of 1
Who is online
In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], John Guajardo and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], John Guajardo and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science