I am trying to determine benzene in gasoline 0%-1% with HPLC UV. My mobile phase is ACN60%-H2040% and my column is a Vydac (Avantor) 250mmx2.1mm, 120A, 5μm. The detector is set at 205nm since benzene absorbs intensively at 205nm and the other compounds such as olefins etc not. Also, the solvent (1:100 sample dilution) is ACN.
I have a very good separation between benzene and toluene, but I notice great RSD in my injections. When I inject the same vial, i have area drift and the area of benzene reduces with every injection. This happens in both samples and standard solutions.
I tried to have the same sample in different vials. Then I notice RSD (4-9%) between 5 vials (not drift with reducing area). Then i tried to have a vial being injected and quickly replace the cap for the next injections. Still the area would reduce with every injection.
Maybe i should check my injector and seal packs, which I haven't yet but,
Do you think there is a problem with my solvent. Has anyone have any solvent issues when the matrix is gasoline?