We are running a 20 minute multi-residue method with almost 500 pesticides on an Agilent 6495C MS and 1290 Infinity II LC system. The method switches between positive and negative mode.
For a while now we have been getting consistently low or no responses for the negative mode compounds, while the positive mode has good response. Running the negative compounds alone, in a method without switching, gives good response. Our technician has replaced the capillary voltage power supply and the high energy diode, neither changed the situation.
Note that we have two other equipment setups of exactly the same type in our lab and neither are experiencing this problem.
We are at a loss for how to continue troubleshooting this. Has anyone seen this before or have any ideas?
We have 6475 and we have encountered similar issues with polarity switching in my lab. Our compound list is relatively short (endocrine disruptors, personal care products in water), the cal curves look awful. When I split the method into 2 separate ones using single polarity, we get good results. Also, when running the polarity switching method, my gain in (-) is at 10 to see the low cal for hormones, while the gain for (+) must be set to 0.2 - 0.4. I placed a service call but so far I am getting nowhere with Agilent. Anyway, I realize I am not very helpful with my response but maybe it would be worth getting their checkout test mix. I don’t have the PN at hand but I am sure it can be found on the Agilent website.