A question towards clarity of your issue - are you worried about the actual ability to integrate, or are you worried about correctly quantifying the amount of substance you are integrating? Optimizing integration parameters is one thing; optimizing your separation is another.
If it's the latter, are you working under USP guidelines/regulations? If so, I found two different numbers, either 1.5 or 2. I think it depends on what the method-specific system suitability parameters are.
http://www.usp.org/pdf/EN/meetings/asMe ... meters.pdf
>1.5 for the critical pair
http://www.fda.gov/downloads/Drugs/Guid ... 134409.pdf
Recommendations:
R, of > 2 between the peak of interest and the closest potential interfering peak (impurity, excipient, degradation product, internal standard, etc.) is desirable.
But, I don't work in the pharma industry, so those who have more knowledge than me will know better.