balance time?
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
2 posts Page 1 of 1
Dear friends...
I am getting this error message from the Refractive Index Detector:
Optical balance movement cannot detect home.
What does it mean? How do I fix this?
Addit info:
software: ChromQuest 4.2
Surveyor RI Plus Detector
Thermo Electron.
Many thanks
it is not a sotware problem but an hardware problem.
something to do with the optics.
most simple RI problems are related to the solvents found in the ref and smp cell.
make sure both are purged with the same solvent.
otherwise it is a more serious device problem