Maximum volume that can be injected for semi-prep

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
I would like to use my C-18 analytic columns (3 x 150 mm), Symmetry and XBridge, for example, for semi-prep purification. Assuming the mass of the analyte is not a problem as its solubility is relatively low, what is the maximum volume that I would be able to inject for such work? What is the formula or rule-of-thumb for such purposes?
Many Thanks.

If your sample diluent = mobile phase, then about 15% of the volume of your narrowest peak. If your diluent is stronger than the mobile phase, the injection volume will have to be reduced, but I prefer to experiment rather than rely on a rule of thumb in that case. If the diluent is weaker than the mobile phase, then you can inject a larger volume; again, rules of thumb are not terribly useful and experimentation is a better approach.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanks Tom for your super-fast reply.
My sample diluent is DMSO and mobile phase is methanol/water/phosphoric acid.
15% for a peak of width of about 1 min, flow-rate of 0.8 mL/min is 0.12 mL or only 120 microliters. That's it?! Ouch.
Wouldn't the injected volume depend on the volume or void volume of the column?

The "15%" rule represents the point at which the injection volume will begin to make a noticeable contribution to the peak width. The column void volume is subsumed into the peak volume (a larger column will be operated at a higher flow rate so the peak volume will be greater)

If you have "excess" resolution between your product and the closest impurity/contaminant, then you can increase the injection volume until the resolution drops below the "adequate" level (whatever that is). I'll repeat what I said at the end of my previous post: rules of thumb are not terribly useful (for this sort of thing) and experimentation is a better approach.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

If your sample is still reasonably soluble in a solvent that contains (much) more water than the mobile phase, you can potentially inject a "ton" of sample. For example, if you can dissolve the sample in a solvent that contains about 20% more water, you can inject about 10 times more sample volume for equal peak distortion (this is a rule of thumb, the actual value depends on the sample). If you have a second pump, you can inject the sample as is, but dilute it just before the column with water or buffer delivered by the second pump. This technique is called at-column dilution, and you can find some information in the literature or at the Waters website. This again permits the injection of a much larger sample volume. How much depends on the composition of your mobile phase etc.

Thanks Tom & Uwe. You've really been tremendously helpful.

My compound of interest is quite hydrophobic and the best solvent that I found for it is DMSO (any added water will only lower its solubility). Now, I do not have an additional pump, which as Uwe calls it, would certainly be helpful. I completely understand Tom that experimentation, and not formulas or rules, is the only way to really see what's going on. But, for a first "guess" or "attempt" may I ask the following: If indeed the closest component (an impurity) is, say, 3 minutes away (before and after), how much volume would I then probably be able to inject?
Many Thanks.

Given that your solvent (DMSO) is signficantly stronger than your mobile phase, your guess would be as good as mine. :cry:
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
7 posts Page 1 of 1

Who is online

In total there are 35 users online :: 1 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Majestic-12 [Bot] and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry