Dionex ICS2000 and legacy AS4A columns
Discussions about IC and related topics
4 posts
Page 1 of 1
Our lab's IC has been running inorganic anions with AS18 columns. We just came across a stash of never-used new-in-box AS4A columns (and the corresponding guard columns) from a prior Dionex system 20 years ago (!), which prompted two questions: 1) are they compatible for anions on the ICS-2000, and 2) will they have survived long storage (always room temp, and still in as-shipped condition)?.
-
- Posts: 65
- Joined: Fri Jul 01, 2011 1:29 pm
Hi
What a luck to find a AS4A; probably in the big blue box? You have to know, the AS4A incl. guard is a so called carbonate selective column. The normal eluent conditions are 1.7mM Na2CO3 / 1.8mM NaHCO3 and isocratic. The AS18 you use now is a hydroxide selective column. I assume therefore that you use an eluent generator with your ICS-2000 which is not compatible with Carbonate eluents. Considering the age and the low capacity of the AS4A (the AS4A was developed late 1980) i would store the column in the museum.
Best, Stefan
What a luck to find a AS4A; probably in the big blue box? You have to know, the AS4A incl. guard is a so called carbonate selective column. The normal eluent conditions are 1.7mM Na2CO3 / 1.8mM NaHCO3 and isocratic. The AS18 you use now is a hydroxide selective column. I assume therefore that you use an eluent generator with your ICS-2000 which is not compatible with Carbonate eluents. Considering the age and the low capacity of the AS4A (the AS4A was developed late 1980) i would store the column in the museum.
Best, Stefan
-
- Posts: 362
- Joined: Tue Aug 31, 2004 6:54 am
Dear ICagain
if you are curious give it a try with the eluent Stephan mentioned. you even may can can change the carb/bicarb ratio. This might show you interesting possibilities of the column.
But for routine work I also wound not change a working system.
if you are curious give it a try with the eluent Stephan mentioned. you even may can can change the carb/bicarb ratio. This might show you interesting possibilities of the column.
But for routine work I also wound not change a working system.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
-
- Posts: 109
- Joined: Wed May 22, 2019 5:19 pm
I checked the ThermoFisher website and they still list them for sale at $1500.
Maybe someone somewhere still uses them. List them on e-bay and see if anyone wants one.For fast, isocratic separation of inorganic anions in environmental waters, the Thermo Scientific™ Dionex™ IonPac™ AS4A-SC column was the preferred column for many years. It was first introduced in 1984 and is the specified column in U.S. EPA Method 300.0 (A). However, the advanced, high-capacity Dionex IonPac AS22 Carbonate Eluent Anion-Exchange column is recommended for fast analysis of inorganic anions in diverse matrices when using carbonate eluents.
4 posts
Page 1 of 1
Who is online
In total there are 188 users online :: 0 registered, 0 hidden and 188 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm
Users browsing this forum: No registered users and 188 guests
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm
Users browsing this forum: No registered users and 188 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science