i just recently had to replace a column after about 2 and a half months of use. It was an agilent DB5-ht. I have been reading a lot about the life of columns hand how you should be getting a certain number of injections out of it, on the order of thousands. However I saw a loss of resolution after 400 samples, probably less than that. I look at supelco bulletin 853b for any problems with the column, i think its a pretty thorough guide.
Anyway i start to get peak tailing, peak broadening and an lowering of the areas. I just keep track of the areas of my internal standards and if they start getting low then i can tell the column is getting worn.
I use MSTFA to derivitize our biodiesel/glycerides, which is a silyating reagent, if what kvaak says is right this could be one reason why my column doesn't last as long, especially if we run samples with high glycerol/glycerides in them. Good to know!