Agilent 1100 autosampler
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
Does anyone know if the screw rods in the Agilent 1100 autosampler which move the transport need greasing or lubrication? I couldn't find anything in the Reference Manual about this. Thanks.
-
- Posts: 762
- Joined: Thu Dec 13, 2007 5:54 am
I have a suspicion that they might. I think it could not hurt to use a little lithium grease on the z- and x-axis rods at the back of the transport assembly. I have done this myself without complication on the z-axis rod. I belive this might avoid some of the 'transport motor overheat' errors that can happen, which usually result in Agilent recommending you replace a $1900 transport assembly.
Time flies like an arrow. Fruit flies like a banana.
-
- Posts: 14
- Joined: Tue Nov 25, 2008 3:21 pm
I was told by an Agilent Service engineer that these rods should not be greased since this will lead to a build up of dust on the rods which will cause errors. He recomended cleaning the rods with a swab soaked in IPA every few months. We have been doing this for several years and have had no problems.
-
- Posts: 762
- Joined: Thu Dec 13, 2007 5:54 am
"He recomended cleaning the rods with a swab soaked in IPA every few months. We have been doing this for several years and have had no problems."
That has to be a pretty annoying task. Do you have to pull the transport assembly out of the autosampler to do this? I can't imagine how you would get to the screw rods any other way.
That has to be a pretty annoying task. Do you have to pull the transport assembly out of the autosampler to do this? I can't imagine how you would get to the screw rods any other way.
Time flies like an arrow. Fruit flies like a banana.
-
- Posts: 405
- Joined: Wed Dec 05, 2007 4:55 pm
I have also heard the warning about lubrication, but don't have any further details. If you have a dusty lab I would worry about things sticking.
I have used the IPA approach to help with a corrosion problem (just soaked a soft cloth with IPA and wiped the rods as well as I could without removing the assembly). Seems to work fine now.
A wipe with a silicone spray lubricant might be a better compromise.
I have used the IPA approach to help with a corrosion problem (just soaked a soft cloth with IPA and wiped the rods as well as I could without removing the assembly). Seems to work fine now.
A wipe with a silicone spray lubricant might be a better compromise.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
ACCTA, Inc.
5 posts
Page 1 of 1
Who is online
In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science