we have a GC method and the details are as mentioned below
Column : DB 624 column
(30 m x 0.32mm ID), 1.8 µm film thickness or Equivalent.)
Injector Temperature : 180°C
Detector : FID
Detector Temperature : 250°C
Flow : 1.7ml/min, constant flow
Carrier gas : Helium/ Nitrogen
Oven Temperature : 60°C to 220°C at the rate of 20 °C per minute and hold at 220°C for 3 min.
Injection type : Split less (Purge flow :- 20 ml/min)
Headspace Parameters:
Vial Oven Temperature : 80°C
Transfer line Temperature: 110°C
Loop Temperature : 90°C
GC Cycle Time : 18 minutes
Vial equilibration Time : 60 minutes
Pressurization time : 2.0 minutes
Loop fill time : 0.2 minutes
Loop Equilibration Time : 0.2 minutes
Injection Time : 0.08 minute
Vial Shake : Low
Loop Size : 1.0 ml
Headspace Vial : 20 mL
Headspace Vial Pressure : 14 psi
Preparation of Standard Solution:
Weigh accurately about 534 mg of Toluene into a 50mL volumetric flask containing about 20 mL of Dimethyl formamide. Dissolve and dilute to volume with Dimethyl formamide and mix well. 5.0 mL of the above solution transferred into a 50 mL volumetric flask and then dilute to volume with dimethyl formamide. 5.0 of the above solution dilute to 100 mL volumetric flask with Dimethyl formamide.
Transfer 5.0 mL of the solution into a headspace vial and seal the vial immediately.
Now the question is that the baseline is very straight and good in th first injection and went bad with other injections , , please can anybody explain how can attach chromatograms also .