What grade of solvents can be used for quantitave analysis in gas chromatography? Can anyone explane to me difference between reference material,analytical standard, just standard for GC and high purity solent? What is acceptable ito use in pharma QC laboratory for quantification?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
Solvents: use a grade of solvent that is sufficiently free fo contaminatns that you do not get interference with analytes. Depending on the analysis you are preforming, you may be able to use reagent grade solvents or you may need specially purified solvents for your application.
A reference material is a material of known composition and of sufficient quality that it can be expected to give repeatable analytical results of a specified value for anyone properly analyzing a sample of the material.
An analytical standard is a material (often a pure compund) that may be used as a reference for calibrating an analytical instrument. In many cases, the standard may be purcachased already disolved in solution, making preparation of working standards fairly easy.
A high purity solvent is one that has been specially purified to avoid particular interferences. Purity has to do with ability to meet specific specificaitons. GC grade solvents have low leversl of compunds that often interfere with chromatograms.
I will let someone from pharma speak to specific issues for the industry, but as a minimum, you need solvents that will not interfere with analytes of interest by either containing compunds that would react with the analytes or otherwise interfere with the measurement. Purity of solvents must be sufficient that results will be reproducable.
Analytical standards should be of known purity, typically 99% pure or better - and reproducable. Variation in standard materials leads to variation in results. In some cases, compunds can not be obtained at high purity, but can be obtained at a certified level of purity. This can work also.
A reference material is a material of known composition and of sufficient quality that it can be expected to give repeatable analytical results of a specified value for anyone properly analyzing a sample of the material.
An analytical standard is a material (often a pure compund) that may be used as a reference for calibrating an analytical instrument. In many cases, the standard may be purcachased already disolved in solution, making preparation of working standards fairly easy.
A high purity solvent is one that has been specially purified to avoid particular interferences. Purity has to do with ability to meet specific specificaitons. GC grade solvents have low leversl of compunds that often interfere with chromatograms.
I will let someone from pharma speak to specific issues for the industry, but as a minimum, you need solvents that will not interfere with analytes of interest by either containing compunds that would react with the analytes or otherwise interfere with the measurement. Purity of solvents must be sufficient that results will be reproducable.
Analytical standards should be of known purity, typically 99% pure or better - and reproducable. Variation in standard materials leads to variation in results. In some cases, compunds can not be obtained at high purity, but can be obtained at a certified level of purity. This can work also.
-
- Posts: 658
- Joined: Thu Jun 19, 2008 11:07 am
Hi
Don covered it well, just adding some pharma specific information from a recent FDA Q and A session:
Residual Solvents in ANDAs:
Questions and Answers (October 28, 2008)
Q11: Would it be acceptable to use a high purity solvent in place of the USP reference standard?
A: Yes, a high purity solvent may be used in lieu of the reference standard if suitable documentation (i.e., certificate of analysis) of the purity and source is provided.
Don covered it well, just adding some pharma specific information from a recent FDA Q and A session:
Residual Solvents in ANDAs:
Questions and Answers (October 28, 2008)
Q11: Would it be acceptable to use a high purity solvent in place of the USP reference standard?
A: Yes, a high purity solvent may be used in lieu of the reference standard if suitable documentation (i.e., certificate of analysis) of the purity and source is provided.
-
- Posts: 2
- Joined: Mon Nov 23, 2009 2:20 pm
Hi,
i've used a standard with 50 VOCs (2000 mg/l ).Then i made 7 standards with certain concentrations 10,25,50,100,200,300 ppb to make the calibration.
The problem is that the peaks in the 10 ppb graph are higher than those in the 25 ppb graph which has no sense..The same happenned with 50 and 25 ppb graphs..
Does anyone know what the problem may be??
(I have already checked the standard solutions and the fibber..)
Thank you!
i've used a standard with 50 VOCs (2000 mg/l ).Then i made 7 standards with certain concentrations 10,25,50,100,200,300 ppb to make the calibration.
The problem is that the peaks in the 10 ppb graph are higher than those in the 25 ppb graph which has no sense..The same happenned with 50 and 25 ppb graphs..
Does anyone know what the problem may be??
(I have already checked the standard solutions and the fibber..)
Thank you!
-
- Posts: 5432
- Joined: Thu Oct 13, 2005 2:29 pm
I have an idea, but I would like to look at the data first. Can you post the calibration graph (instructions in a sticky at the top of the LC page) or give us the raw data.
Peter
Peter
Peter Apps
-
- Posts: 2
- Joined: Mon Nov 23, 2009 2:20 pm
what exactly do you mean by saying "raw data"..??
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
retention times and most importantly, the area counts of the peaks in question.
Rod
Rod
-
- Posts: 5432
- Joined: Thu Oct 13, 2005 2:29 pm
[quote="dimxaz"]Hi,
The problem is that the peaks in the 10 ppb graph are higher than those in the 25 ppb graph which has no sense..The same happenned with 50 and 25 ppb graphs..
quote]
So: 10 ppb is higher than 25 ppb, and 25 ppb is higher than 50 ppb, therefore 10 ppb is higher than 50 ppb.
Are the areas for 10 ppb the same as those for 100 ppb by any chance ?
The raw data that I need to see is the peak areas for each level of standard (i.e. each concentration). To save my time, I would far rather see the calibration plots.
Peter
The problem is that the peaks in the 10 ppb graph are higher than those in the 25 ppb graph which has no sense..The same happenned with 50 and 25 ppb graphs..
quote]
So: 10 ppb is higher than 25 ppb, and 25 ppb is higher than 50 ppb, therefore 10 ppb is higher than 50 ppb.
Are the areas for 10 ppb the same as those for 100 ppb by any chance ?
The raw data that I need to see is the peak areas for each level of standard (i.e. each concentration). To save my time, I would far rather see the calibration plots.
Peter
Peter Apps
8 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science