migration time shifting in chiral CE assay

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi, All,
I am runing a chiral CE assay. The condition is roughly as follow:
50umx56cm capillary, 5% sulfatd CD with 10% acetontrile in 25mM phosphate TEA buffer, pH 2.5, -15kV, and the peaks come out around 20min. The migration time of all peak are roughly decreased about 0.2 min after every injection (5 min rinse with buffer between injections). If no acetontrile in buffer, there is no shift. Is this due to acetontrile evaporazition or something else? The solvent evaporazition seems fairly quick to me if it is the reason.

Thanks a lot.
Hi, All,
I am runing a chiral CE assay. The condition is roughly as follow:
50umx56cm capillary, 5% sulfatd CD with 10% acetontrile in 25mM phosphate TEA buffer, pH 2.5, -15kV, and the peaks come out around 20min. The migration time of all peak are roughly decreased about 0.2 min after every injection (5 min rinse with buffer between injections). If no acetontrile in buffer, there is no shift. Is this due to acetontrile evaporazition or something else? The solvent evaporazition seems fairly quick to me if it is the reason.

Thanks a lot.
Maybe try first after every run the following: Rinse 1 min. with 0.1M NaOH (to reactivate and clean the inner surface) then rinse 1 min. with water and finally rinse at least 3 min. with buffer. I would think that this solves your problem.

Try also putting some extra run buffer vials on the system and running each injection with a pair of fresh buffer vials to see if that removes the MT drift between runs. I am wondering if you get a pH/electrolyte change during the run which is somehow accentuated when the buffer contains ACN.
regards, Rob
____________________________________
Rob Sample Consulting - working worldwide to bring analytical expertise to your laboratory.

Bob and Fiz,
Thanks for the suggestions. I used 0.1N NaOH rinse, and it gave the same result, ~ 0. 2 min decrease every injection. Using new buffer vials helps, but there is not very convenient in routine analysis.
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry