The problem is that water has a very high degree of thermal expansion causing a small "explosion" inside the injection port as it vaporizes resulting in bad peak shape.
I’d say it’s a question of injection volume, as well as injection port temperature. I'm not a GC expert but I’ve seen people injecting mixtures containing up to 80 % water.
Danko: Most of the time, the concentration of chemical is determined in the aqueous phase and the remainder is assumed to be sorbed to the soil. A better procedure would be to extract the soil with solvent after removing the aqueous phase so that both the aqueous phase and sorbed analyte can be determined.
If I were going to use something else than water for a sample solvent (e.g. ethanol, methanol, or whatever), I would certainly extract my analyte in that solvent from the beginning. Also, here one needs to consider the solubility of the compound of interest. Is it equally soluble in water and in the organic solvent? What if the compound stays adsorbed on the soil particles and never dissolves in the water?
I was just trying to emphasise the need for careful planning of the exercise in question.