Operational Qualification in HP 1100

Posted: Sat Dec 11, 2004 8:01 pm
by Tequila
Dear members of this forum,

My question is about the verification (operational qualification) of Liquid Chromatographer from Agilent Model 1100. In this step I used the new restriction capillary and the main difference with the old is the length, it is ten times higher. When I turned on the pump, the pressure increase a lot and I turned off inmediatelly. I want to know if anyone have experience in this step of verification (composition gradient) and until what maximum value of pressure can work without damaging the pump. This information will be very valuable for me.
Thanks in advance for your help,

Diego Delmonte

OQ of your HPLC system

Posted: Mon Dec 13, 2004 11:14 am
by Antje
Dear Diego,
we use for OQ of our HPLC 1100 (Agilent) the same capillary, wich is used by Agilent: Part-No. 5022-2159. It works very well. Wich part no. do you youse?
The backpressure of the system can go up to 400 bar. This performs the leak test also.
I hope, this will help you!
Best regards!
Antje