We've recently acquired a Varian NatGasB instrument from another lab and are currently configuring it for permanent gas analysis.
The NatGasB instrument is a CP-3800 instrument purpose built for the analysis of natural gas. It's a fairly complicated instrument so I'm only using it in it's most basic form just now.
The separation of oxygen, nitrogen, methane and carbon monoxide is handled by a Molseive 13X 45/60 mesh column which is then backflushed.
The instrument runs isothermally at 120 Deg C.
Initially the separation was good

I then switched all 3 250ul sample loops for 100ul sample loops to improve peak shape at higher concentrations and then got this.

Over the last few days the resolution has degraded further to this

No additional work has been done to the instrument during this time.
The plan for today is to reinstall the original sample loops and try and get it back to where it was when I first got it.
Regards
Rich