I am a new GC user working with an old HP 5890 Series II with FID and TCD. We have a single injection port, and split the gas into two columns leading to the two detectors. This set up worked well for our mixed gas of hydrogen and chlorinated hydrocarbons, until one day, right under my nose, I saw the hydrogen signal disappeared on the TCD, together with the FID peaks.
I performed some trouble shooting procedures, but haven't confirmed the problem yet.
1. The FID and TCD still give non-zero, non-maximum signals (base lines). The baselines carry certain normal noises, and can jump up and down when the flow rate is changed.
2. I did not detect any leaks in the system. Flow rates look like normal.
3. While blocking the inlet purge line at the injection port, and injecting several micro liters of liquid methanol, the detectors still can not pick up any significant peak.
Overall, it looks like either nothing injected got into the column, or the TCD and FID lost sensitivity at the same time. It is a strange thing to me. Can anybody provide some insights? Thanks a lot!