I have a problem with RP-HPLC purification of oligonucleotides.
I am using a three moths old 250/4.6 Nucleosil 300-5 C18 column to purify oligonucleotides with triethylammonium acetate / Acetonitrile buffer system, and I wanted to regenerate the column.
In a first attempt I reversed the flow on the column and applied a 1 mL/min flow with the following solvents: Acetonitrile, MeOH, THF, MeOH and back to acetonitrile. Every wash lasted 25 minutes. The result was that I eluted a lot of material from the column; the separation quality increased, but the back-pressure raised from 120 bars to 160 bars.
In a second attempt I used 0.4 mL/min and only Acetonitrile,THF and back to acetonitrile, and this time the pressure raised evem more to 180 bars.
I don't know what I am doing wrong.
Does anybody of you know how I could wash the column and still maintain moderatly-high pressure? Do you think I will be able to recover the column, or should I trash it? Would you suggest to use ionexchange instead?
Thank you for your help
Best regards,
Giulio