danko » Thu Oct 08, 2009 1:10 pm
													
 
					
						Hi Lotusx06,
The answer to your question is actually quite simple: Scale up!
More detail follows:
1. Achieve the desired separation on an analytical column.
2. Determine the load-ability limit on that same column, by loading increasing amounts, until you observe considerable chromatography deterioration.
3. Do some calculations – especially in the context of how much stuff you want to purify (at the time/per batch)
4. Contact the column manufacturer and ask for the column format you determined you needed from the calculation work, or something similar.
5. And you’re in business.
Best Regards