PeterB » Tue May 10, 2011 2:59 am
You have an equilibrium (for your 1,4-Dioxane) between your water and your headspace, and an equilibrium between your headspace and the SPME fibre. Using the labelled Dioxane as your Internal Standard will minimise the effects changing those equilibriums will cause, but ideally, you should keep them as consistent as possible, particularly if you're trying to keep your abundances consistent. There really isn't any need for them to be consistent in an isotope dilution method, but it sure helps with diagnostics if things go wrong.
So, for calibrations and samples, use the same volume of water, the same amount of salt, and, if your samples are fairly consistent in ethanol content, about the same amount of ethanol. You want the partition between the water and the gas to be as favorable as possible (hence the salt to push the dioxane out of the water), and also want a good interaction between the headspace and the SPME fibre. Ethanol will increase the solubility of the dioxane making it harder for you to recover (hence your lower abundances), but it should equally affect your Internal Standard, so you should get the right result providing you're getting a decent amount of the dioxane onto your fibre.