Shimadzu GC-17A: how far to insert column into ECD?

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
How far should a capillary column be inserted into an ECD detector on a Shimadzu GC-17A?

I'd RTFM, if I had one.
Michael J. Freeman
Belle Chasse, LA

76mm. You should have a measuring jig marked with an E for this.

76mm. You should have a measuring jig marked with an E for this.
Thanks. I'm thinking about ordering a "Capillary Installation Gauge." Stupidly expensive, but sometimes the right tool is worth it.