Sample loop in Waters 717plus autosampler
Posted: Mon Sep 14, 2009 8:55 am
by chhubert
Dear all,
I want to change the sample loop of the Waters 717plus autosampler to 100uL in order to minimize the system dead volume. Is there any one who has such practical experience ? For Waters 717plus autosampler, there are only two sizes of sample loop - 200 and 2000uL. Thanks a lot.
Hubert
Posted: Mon Sep 14, 2009 11:18 am
by danko
Just remove the existing loop (200 µL ?) and then install the 100 µL loop (the fittings are the same)
Best Regards
Posted: Mon Sep 14, 2009 1:09 pm
by AA
Simply replacing the loop will not reduce the volume enough to make a significant impact on the chromatography. The option is to install the Low Volume Stainless Steel Tubing Kit [WAT055886] for the 717.
Posted: Mon Sep 14, 2009 1:26 pm
by chhubert
AA: I have replaced the low volume stainless steel tubing kit already. However, I find that it only reduces the system dead volume between the pump and the autosampler but not the autosampler (or sample loop) to column.
Danko: In the front panel of 717plus, I can only select the sample loop of 200uL or 2000uL. Will there be any problem (instrument-wise) if I change the loop physically to 100uL but keep the sample loop setting on the panel to 200uL?
Thanks a lot.
Hubert
Posted: Mon Sep 14, 2009 1:37 pm
by danko
No problem at all. Just choose 200 µL. It works perfectly with 100 µL loop.
Best Regards
Posted: Tue Sep 15, 2009 4:16 am
by chhubert
Thanks a lot, Danko. I'd try and see how it improves my work.
Hubert