Hi all!
Does anyone have informations to give me about the "hidden" problems of using Empower 2 to pilot HPLC Agilent 1100? For examples, what fonctionnality do we loose?
What kind of problem do you have for using the HPLC in routine and how do you solve it?
I'm actually doing a pilot test and trying to find what problem we can expect in the futur to solve them as quick as prossible.

Thanks a lot!

Rico

You "loose" all diagnostics and tests, including autosampler adjustment/parts exchange functions (arm movement for needle excange and others), lamp intensity/cell tests.....

Many "finetune" settings are not accesible from Empower - like how deep needle goes into vial, special injection functions, different compressibility coefficients for different channels in pump.....
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry