Oxygen headspace method driving me crazy...help plz

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
hi all... i'm an analyst with a relatively new experience in GC...i have a method for the analysis of oxygen in the headspace of vials and ampoules...i have agilent 7890A using a molecular seive column 30m x 320mm x 12um...my method uses helium as carrier gas at a flow of 1ml/min with inlet temp of 250C witha a split ratio of 4:1 and the sample is injected manually, column oven at 60C, TCD at 300C...everything was going just perfect...an example run is below:
Image
today i had some samples to analyze, after conditioning the column i injected a sample and i had the following chromatogram:
Image
i tried many injections with the same result, i removed the column and cut from its two ends and re installed it and no change occured, i changed the inlet septum with no use...
please help me what could be the possible cause of this change in one day and what can i do?

Dear Dr.Omar,
It seems by removing the column out of gas and cutting both extremities, you contaminated the column with air and specially moisture. Some moisture is now traps in the column and it is why your system baseline is so noisy.
I would suggest to regenerate(bakeout) the column at 300Celcius on 12 hours period with purified helium as purging gas at a minimum of 10cc/min.
Note : make sure you didn't create leaks on the fittings you played with. Loose fittings or bad installed fittings could also cause such bad reading.

I already have seen such situations with TCD.

regards
Dany

I would guess that your syringe is blocked, or you have a serious leak around your injector.

In any case your flow rate is incorrect, or you are not injecting any sample onto the column due to a blocked syringe.

Since your detector 'zero' is higher now than before you probably have a leak in your injector somewhere causing contaminated carrier gas at the detector.

Trouble shoot the hardware and you should find the problem.

best wishes,

Rodney George
CONSULTANT

Thanks alot for your replies...i will try the approaches you suggested
4 posts Page 1 of 1

Who is online

In total there are 87 users online :: 1 registered, 0 hidden and 86 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Majestic-12 [Bot] and 86 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry