peak tailing(ergent!)

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
Dear all,

Today I performed our LC periodical verfication and there may be something wrong in the system: the caffeine standard solution injected the different systems(the column is removed and connect the injector and detector with a capillary tube) and the results confused me(the chromatograms as following):

Image

The upper(Aug. 26) and the middle(Aug. 06) are generated from the same LC(LC-1) , and the bottom chromatogram is made in another LC system today. With the same constant conditions, the chromatograms lead to a big problem, and I don't known what I do. Can anyone help me? Thanks in advance!
The God had ever have three apples. Adam was tricked eating up one of them in Eden, and the second dropped from the tree and hit Issac Newton on his head, then what happened on the third one?Interestingly it was bit by Steve Jobs!

It looks like you have injected much, much more in chrom No 1.
Would you tell us how high the Y-scale goes (for all 3 chroms)?

Best Regards
Learn Innovate and Share

Dancho Dikov

check - as suggested by danko
clean if needed (purge everything)
repeat and report :)

Thank you! Danko & Grzesiek

The Y-axis is 10 and the area is 30 about, and I the method is checked many times so I sure that the injection amount is OK, because the same standards are annalyzed on LC-2 system under the same condition showing good peak shape and perfect linearity. I think the system 1 is not purged thoroughly and some unknown materials tails the peak(the system 2 was purged thoroughly), but I don't know what component-pump head, injector,tube & line or detector- leads the tailing(because the column is removed, so the chemi-absorption on column is unbelievable), now I doubt the purging is the main issue, OK, I'll purge the system before re-analysis tomrrow.

Good luck to me! :P
The God had ever have three apples. Adam was tricked eating up one of them in Eden, and the second dropped from the tree and hit Issac Newton on his head, then what happened on the third one?Interestingly it was bit by Steve Jobs!

If you injected the same amount in all, it looks like a drastic flow change occurred in the first chromatogram where the downslope decreases.

This statement is confusing: " (because the column is removed, so the chemi-absorption on column is unbelievable), "

I suggest some possible causes could be either the autosampler loop has not been purged of a previous mobile phase by a couple of full volume injections or, less likely, that there were large voids in the column replacement connections.

Please keep having fun,

Bruce Hamilton

Hi Austin,
Correct me if I’m wrong but from my point of view the first chromatogram’s Y-axis represents double digit numbers, whilst the second and the third chrom’s. Y-axes are single digit scales.

Apart from that, I think Bruce’ suggestion;
[quote]“the autosampler loop has not been purged of a previous mobile phaseâ€
Learn Innovate and Share

Dancho Dikov
7 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry