Hi people!

I work at a GC6890/MSD5973 (with a turbo pump and cuadrupole).
The column is a HP 5ms. The flow is 0.8, He.

I have a problem.

The system worked good 2 weeks ago. Then, I shutted off the system and installed a UPS. After that, the system's sensivity fall a lot.

The autotune is OK (EMV 1700, Repeller 31)
I changed septum and liner.
The gold seal was changed 1 month ago.

What can the problem be?

Thanks

YOu indicate that autotune is 1700 V. What was it before? (If it has not changed, the mass spec is fine and you need to look for a chroamtography issue - like column installation or inlet issues.)

How is the background air? (Leaks kill sensitivity)

Also, what are you using to measure the sensitivity of the gc/ms system?

If you did not do anything except shutting down the instrumen (to put in the UPS) and the sensitivity reduced a lot, the thing may be that you did the shutdown not properly. Or if you have reinstalled the column (in the MS side), that could be the culprit.
One more thing: you shoud check the groundness of the lines after the UPS.

YOu indicate that autotune is 1700 V. What was it before? (If it has not changed, the mass spec is fine and you need to look for a chroamtography issue - like column installation or inlet issues.)

How is the background air? (Leaks kill sensitivity)

Also, what are you using to measure the sensitivity of the gc/ms system?
I use the usual standard, and the abundances falled 90%

I used the usual standard, and the abundances falled 90%.

Because the 5973 adjusts the detector voltage to adjust for sensitivity changes in the mass spectrometer, I would say that the problem is chromatographic. Check the connections you made while doing inlet maintenance to be sure that you do not have a leak. (A leak can result in loss of sample by splitting flow between the column and the outside of the instrument.) Also be sure that you are using a good syringe. While it does not happen often they can wear enough that the sample can blow back past the plunger.

Also, be sure of the standard. Some do not keep well and sometimes standards that keep well are accidentally spoiled.
6 posts Page 1 of 1

Who is online

In total there are 184 users online :: 1 registered, 0 hidden and 183 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 183 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry