Jasco UV detector 1575 Range setting?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
Does anyone know what range setting "S" (short) means on the display of a Jasco intelligent UV-VIS detector model 1575? There are a number of possible numerical 'Range' settings, increasing 2-fold from 0.0005 up to 2.56. Presumably these are sensitivity settings of Absorbance units per 10mV. I can't figure out what "S" represents.
-
- Posts: 56
- Joined: Sun Jun 22, 2008 9:24 am
The S setting stands for Short, and is to be used for test purposes only.