This is in regards to a previous post, but the issue still has not been resolved if anyone has any other idea pls let me know.
Shimadzu GC17A MS QP5000
Below is a little history:
Shimadzu PM 3/3/09
Terra Tech PT PM 3/11/09
In the 2nd week of May we started to lose sensitivity, the ISTD area counts went from 1,000,000 down to 200,000. I made new solutions/recalibrated and this did not improve the results. Tune file showed a slightly dirty system, peak 502 jagged. I started tech support with Shimadzu:
6/8 methanol rinses and system bake out, no improvement
6/9 complete leak check, small leak found, tightened fitting, no improvement.
6/9 direct injection results not any better than PT injections
6/10 changed He tank, no improvement
6/15 cleaned ion box, slight improvement
6/22 changed out inlet parts/consumables, no improvement
6/30 replaced the column, saw immediate improvement, but only lasted a few days and returned to initial state
7/6 contacted EST made some pressure adjustments, saw improvement, but again didn’t last.
7/10-7/19 on vacation
7/21 cleaned lens stack, noticed it was installed backwards, from PM in March corrected issue, saw immediate tune improvements, but sample runs have not improved
7/24 installed new traps in PT, no improvement
I have done direct injections, no improvment. I also recently had a Shimadzu Tech in here and he installed a different column and it passed specs. We reinstalled columns that I had and after multiple runs it returned to a lower sensitivity. We have not tried multiple runs on a different type of column due to time constraints. Another contract lab said our quads needed to be re dipped and the analyzer replaced, however the tunes look fine. The tech is going to return next week to look at the isssue again if anayone has any ideas I could pass along to him that would be great.