I have a problem with GC ms varian, the report/ratio 19/18 is between 20-30% after bekout but after one day it believes at 47% and it is established at 60-67 % although I do not have any external leak.
The conditions of the instrument are:
external source, damping gaz: 1 ml/mn, T gater: 425°C, T source 200, T trappe: 200, T tranfert line: 300, T manifold: 40, T colonne: 200, T inj: 230,
split 1/10, pump foreline: 35 mtorr, pump turbo: 16,1 µtorr, current: 241 mA,
pump speed 100%.
when i test the air water check, i obtain the result:
base peak: 28, amount: 200
32 is the 1/4 of 28
the width of 28 is 0,8
I used the He "N 60" (purity:99,99999)
I worked with another bottle of He, i changed the filter but I board obtained the same results.
when i make the autotune, the result is: test autotune passed. the ration of 19/18 believe down 50%, but after a few hours it believe at 60-67%.
when I make the test of HMN, the base amount is 5.
Note: in the software, the report 19/18 is acceptable when the value is down the 50%
Please can you help me.