i injected also pure acetonitril ( only acetonitril 100% ) i do see the same peak and if you see the word file you will see that the peak appears ,,, i am really frustrated for more than six months now with this proplem !!Can you give us a link to a chromatogram of a methanol blank? Do you see the same peak there? If not, then the peak is *not* coming from the methanol. You have some other source of contamination which you will have to track down (exposure to other solvent vapors, perhaps??).
[size=12]please see the word file i attached now
http://www.2shared.com/file/7421405/c0c ... vent_.html
,the methanol blank contains the contaminant peak [/size]
If you *do* see the peak with a methanol blank, try injecting a "blank" of 50/50 methanol/water. Assuming you use the same bottle of methanol to prepare the mobile phase and the blank, you should then see a negative peak at that same retention time. If you *don't* see a negative peak, then, again, your contaminant is *not* coming from the methanol.
i do inject 50 50 % methanol water i saw a negative peak ( the peak down )
If you *do* get a negative peak, try injecting degassed methanol. Dissolved O2 in methanol does have some absorbance at 210nm.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.