I have an Agilent 6890 GC with an Agilent G1888A headspace sampler. Recently I changed the G1888 from manual pressure control (MPC) to electronic pressure control (EPC), for the carrier only. That involves cutting the carrier line and redirecting the flow through the HSS so that the only carrier source for the inlet is through the HSS transfer line. I connected the transfer line close to the inlet and wrapped the exposed section of tubing with glass wool and aluminum foil to minimize any cool zone going into the inlet. I don't use an autosampler tower on that inlet.
There is one method which heavily dominates the use of the HSS right now where an inorganic salt with a highly oxidized organic ligand is digested in 44% w/w phosphoric acid and the solution in phosphoric acid analyzed by headspace GC. This method continues to work well with no surprizes after the change.
However, I started working on a new method using DMF as the diluent, and I started seeing something weird, which I can't remember ever seeing before I changed the instrument.
Simple background on the method - uses a 60m x 0.54mm x 1um Phenomenex ZB-WAXplus column and starts out on a gantle ramp from 40C to 60C over ten minutes. The linear velocity is held constant at 20 cm/sec. Right now the HSS zone temperatures are oven 135C, loop 155C and transfer line 165C, but the same thing is seen at lower temperatures. Inlet temperature is 180C.
When I first inject a vial of DMF (20 mL vial, 1 mL in the vial), I get a monstrous peak at just slightly later than five minutes. This peak tails badly even though no other peaks are showing signs of tailing except TEA (even the DMF peak looks pretty good). Note that for a 60 m column at 20 cm/sec, this peak is pretty much unretained. With successive injections of DMF, the peak becomes smaller and smaller until it reaches a quite manageble size which doesn't interfere with anything else I'm looking for. It always tails. Even after leaving the instrument idle for a while the area never goes up again, as long as the instrument isn't used for other method in the meantime.
If it's a real thing I figure it would almost have to be a light hydrocarbon, because even heptanes are retained on the column, and anything even slightly polar would probably be quite well retained on that wax. That doesn't make any sense at all to me, though.
I put my work down to go on vacation and many more runs of the method with the phosphoric acid diluent we done. When I came back I took the HSS back again to work on my method with the DMF diluent, and lo and behold, the problem is back, with the exact same signature, a giant apparently unretained peak rapidly dimishing as injections go on. Swapping out the inlet liner has no effect. This is a split method, I'm using a Phenomenex split/splitless liner with glass wool.
Does this ring any bells for anyone? I'm also pursuing it with Agilent to see if they have any ideas.
Thanks!
Stephen