How can we define the peak width of a flat-top peak? Is there any equation or program? Should we use peak width at half maximum or baseline width?

If you have a flat-topped peak, most likely your detector is overloaded and you are above its maximum signal range. In this case measuring peak area accurately is impossible.

To be more precise, you could determine the area of a truncated peak, but that would not be the true area produced from that quantity of analyte. So even if you could determine the area of a truncated peak, your calculated value of the quantitative result would be incorrect.

GC software will probably not work well if you ask it to give you width at half height of a flat-topped peak but I do not anticipate any problems if you ask for baseline width. The problem with baseline width is that it can be sensitive to minor changes in peak shape unless you set the integration parameters to give robust detection of peak start and end.

I'm interested in why you are working with flat-topped peaks ?

Peter
Peter Apps

Why do you need to know the width?
And how accurately do you need to know it?

As others have pointed out, if you have a flat-top peak, you are well outside the linear range of your system. Width at half-height is meaningless because you have no way of knowing the true maximum height (your peak is truncated). Width at baseline is determined by drawing tangents to the inflection points. With a truncated peak, you cannot accurately identify those inflection points. That said, a manual measurement with a pencil and straight edge can give you a crude estimate. Depending on your purpose, that may be good enough.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Hi!

If you are looking at fluorescence (FLD?) there sometimes are what the manufaturers (Waters) call "sticky settings" to use when setting up a new method. One of these settings is for example pmt/100, pmt/1000 or pmt left normal. These settings make it possible to see saturated peaks as if they were not saturated, the FLD is somewhat less sensitive.

Could this perhaps help you out?
/Mattias
6 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry