Chemstation tailing factor determination

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
I was wondering if anyone could tell me if the peak height used for determining peak width can be adjusted in Chemstation. I am working with B.03.02 and need to get some alternate tailing factor values.

It appears to me that the calculation is fixed at a height of 5%, but I was hoping for some verification before I move on. Thank you for your time!

Sean

Every "width" parameter in reports is the peak width at half height. The only Tailing factor calculation uses the USP definition. There is a "symmetry" parameter, but it is based on statistical moments (I think) and does not generate a commonly useful number.

There are some additional values calculated in the Performance report styles, and they might help you.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
2 posts Page 1 of 1

Who is online

In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry