Hello Guys:

I have less experince in using CI-GC-MS in negative mode. Currently , I met a strange problem. When I repeat to inject a same standard solution. The signal intensity was kepting on decreasing(significantly). After I changed the ion volume, the signal intenisty of first injection went back to the higest level. However, the signal intensity of following injections was repeated to keep on decreasing ( significanrly after each injection). Any suggestions I should check where this kind of problm comes from?

Thank you so much

Did you happen to clean the source recently?

Typically when I clean our CI source it takes quite a few injections for the system to equilibrate and decent integrity to return.

Also how's your PCI 19 m/z peak? Too much water can result in significant decreases in area counts between runs.
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: deevencoo9 and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry