I'm purifying a natural ligand. I expect it to be ~35-55 amino acids.

Following a preliminary bulk C4 solid-phase extraction step, the first HPLC step is a C4 10 mmID column. The gradient is 27-50% CH3CN over 60 minutes (0.1% TFA). Ambient temperature. I loaded 15 mg extract (Vydac catalog claims the column can handle 50 mg). The active fraction range is Gaussian, though has some tailing. It spans t=14 min to t=40 minutes. In other words, it spans over 1/3 the chromatogram.

At first I would suspect something may be wrong with the column, however, a separate ligand we are screening for elutes AFTER the broad peaks and spans only 3-4 fractions (minutes) according to our assay.

What is the possibility that, in spite of the broad active fraction range for the first, there is somehow only one active species?

Just an idea: Maybe you have (or would have) many peaks/compounds, but inadequate selectivity thus no actual separation is causing the described “mountainâ€
Learn Innovate and Share

Dancho Dikov
2 posts Page 1 of 1

Who is online

In total there are 85 users online :: 1 registered, 0 hidden and 84 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 84 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry