Hello all; I have to develop a quantitative method for volative compunds from a natural product. I am thinkng to use SPME; but I allways has been concern about pressition with HS SPME; in this case I thionk to start do ing some test with diferent times, fibers and temperature; bur I wonder about the best quantitative method; should be extarnal STD fine? or must I go to Internal STD?
Have any one an advices??

Thanks in advance and Best Regards.

Oscar.

Dear Oscarbal,

There is no simple answer for your question. Generally, many SPME quantitative methods use external calibration. But here serious precautions should be taken regarding the complexity of the sample matrix being analyzed. If the samples are "clean" (in other words not too much complex) the external calibration can be used.
But if you are dealing with complex samples in which interactions between analyte(s) and matrix compounds can not be ruled out, you should go with internal calibration or standard addition methods. When using internal standard calibration, you assume that the internal standard added to your standards and samples behaves similarly as your target analyte(s). This is very important during SPME because this sample preparation technique is equilibrium method per se, so if you're using internal standardization you should choose internal standard that will be similar to your analyte (sometimes isotopically labeled standards are recommended, as they have nearly the same distribution constants to SPME fiber coating as the original analyte).
Ok, you're planning to use headspace-SPME so generally speaking the headspace is "always" less complex than the liquid sample under. But don't forget that the possible processes in the liquid sample (association, complexation, partitioning of analyte to the matrix compounds or phases) can alter the liquid-gas equilibrium for the target analyte and hence will change its amount in the gas phase which is readily available for extraction (partitioning) onto the SPME fiber. Essential oils are compex mixtures of different compounds (monoterpenes, sesqui- or di- terpenes, volatile aromatics, lipid components, phenols...), so if you SPME an essential oil in water you'll probably achieve decent quatitation using external calibration, although the internal calibration is better choice here, I think (especially if you could obtain isotopically labeled internal standards for your analytes).

Did you search the scientific journals and find anything similar to your situation? Are you planning to extract neat essential oil or you're planning to do direct headspace extraction from a raw plant material?

Best regards

Headspace SPME will only extract the most volatile, non-polar compounds from the food matrix you want to analyze for volatile compounds, and thus give a partial picture of the volatile aroma profile.

I might suggest using Stir Bar Sorptive Extraction (SBSE), a method similar to SPME, but far more sensitive and reproducible than SPME, as well as more durable than the fragile SPME fibres. SBSE utilizes a magnetic stir bar coated with polydimethylsiloxane coating (or other more polar films), where you would place a stir bar in an aqueous solution with your sample and stir for a set period of time, then thermally desorb the extracted volatile organic compounds onto the GC column for analysis.

Because of the greater amount of PDMS film and surface area than SPME, a greater concentration and selection of compounds may be absorbed onto the stir bar. Since PDMS is non polar it has low affinity with water, and will only extract the more hydrophobic volatiles. This means the partition coefficient (Log P for octanol:water) of the volatiles will determine % recovery for each compound. Thus, samples with high fats or alcohol must be diluted to under 2-3% fat or 10% ethanol. Because of the greater amount of PDMS film, they say an increase of up to 1000 times sensitivity can be achieved as well as being able to extract far more polar compounds. The graph below shows how compounds with much lower partition coefficients can be extracted. In my experience, I have seen small peaks for ethanol and acetic acid, both very polar compounds in wine samples.

Image

As for quantitation, you will have to use internal standards in all sample solutions and compare these internal standards to internal standards in a standard solution with known amounts of the compounds you wish to be able to quantitate. Since GC uses very small amounts of sample (unlike HPLC where a fixed volume loop lends well to external standard curves), internal standards must be used to compensate for instrument and extraction (sample prep) variation. It must be empphasized that since the absorption of each compound to the PDMS coating is an equilibrium effect, different recoveries for each compound will be noticed (same as SPME), and therefore all compounds you want to quantitate should be in the standard solution at known quantities and this solution can be run with several dilutions to obtain a concentration curve, once normalized using the IS. Although it is theoretically possible (but not recommended) to estimate the concentration of compounds' recovery (if compounds not commercially available) based on their Log P and similarity to other quantitated molecules used in the standard solution.

I would suggest gently comminuting your sample and adding deionized RO water and 20% NaCl (this aids in extracting the more polar volatiles), and stirring with stir bar for 1 hour (1000 rpm). Then use a thermal desorption unit to release volatiles from stir bar onto GC and run analysis. If you do not have access to a thermal desorption unit, some researchers have actually inserted PDMS stir bar into GC insert liner before run, and then thermally desorbing onto column at high injection port temp concentrating onto column using low oven temp before GC analysis. Another option, although I am not very familiar with this aspect, is to use an organic solvent to desorb extracted volatile compounds from SBSE stir bar and then do a straight injection onto GC column.

These links will provide you with more information:

http://chromatographyonline.findanalyti ... ail/599227

http://www.gerstel.com/en/twister-stir- ... action.htm

Thanks a lot Zokitano and Carvone;
certainly SBE looks quite interesting but in this case I feel more confortable with SPME due is a technique I has used before; but my problem I allways used SPME in waters sambple and HS for qualitative analysis and I do not know what is the best approach for for quantitative analysis of essential oils in Black ppeper. I was looking for some scientific articles; but like always I can find the abstracts but is a kind of hard for me here to get these papers. I am using FID so I believe isotopic compound is not an option; i wonder is someone knows a good Internal standard that I can use; in this case I have to analyze arround 15 diferent compound but i do not think to use one Internal STD for any analyte; any suggestion or help is wellcome.

Thanks and Best Regards.

Oscar.

Hi Oscar,

You may want to purchase a black pepper essential oil, and run a straight injection of the essential oil on the GC or run a straight injection of the essential oil diluted in a solvent. This will be much cheaper than purchasing the 15 individual compounds, and allow quantitation provided all 15 compounds are also in the black pepper essential oil. Then run the GC with a headspace SPME of your sample followed by a Headspace SPME-GC run using standard addition of the black pepper essential oil to your sample. This will allow you to quantitate based on calculating a SPME recovery factor.

As for an IS, select a compound that is not at all present in the black pepper essential oil and has a good recovery in headspace SPME, but it shouldn't matter otherwise, since you will be calculating quantities based on the standard addition on the same compounds common in the black pepper essential oil and the IS is just a normalizing factor for each chromatogram. Most important is to be consistent in your sample prep methodology for reproducibility. Even comminuting your sample too severely will release volatiles and increase your variation.

The following is freely available and apparently uses SPME ( I haven't downloaded it ). I'm sure there are many others, it was just the first on a Google search.

However, you should still go to the library and obtain some of the many other papers on black pepper oil GC analysis. There are even ISO standards for the commerial product. Why reinvent the wheel?.

Microwave Extraction of Essential Oils (From Black Pepper and Coriander) at 2.46 GHz. Balasundaram Ramanadhan

http://library2.usask.ca/theses/availab ... 05-135622/

University of Saskatchewan Library Electronic Theses & Dissertations

Bruce Hamilton
6 posts Page 1 of 1

Who is online

In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry