Measuring Autosampler Carryover
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
A lot of HPLC manufactures are now claiming ultra low carryovers of less then 0.005%. Yet they refuse to publish the method used and the actual data to back it up. Is there are standard method using a appropriate analyte for measuring carryover at this level?[/b]
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Carry-over is not a standard/universal test that is applicable to every situation. A system could exhibit zero carry-over with a certain compound/analyte, injector-wash-liquid, washing procedure (e.g. time or volume) eluent, injection volume etc. etc.
So…you’ll have to design your own test procedure that is relevant to your needs (e.g. relevant compound etc.)
Here is a kind of starting point that you can build up on:
1. Remove/shortcut the column.
2. Set the flow on.
3. Inject a relevant volume of the sample solvent that you use. Might be several times, until you see no peak at the sensitivity level you’re interested in.
4. Then inject your sample solution and record the peak area.
5. Inject again the sample solvent and record the peak area – if any.
6. Divide the peak area of the solvent/blank peak by the peak area of the sample peak and multiply by 100.
7. The resultant value is percentage carry-over that is relevant to this particular method of analysis.
Best Regards
So…you’ll have to design your own test procedure that is relevant to your needs (e.g. relevant compound etc.)
Here is a kind of starting point that you can build up on:
1. Remove/shortcut the column.
2. Set the flow on.
3. Inject a relevant volume of the sample solvent that you use. Might be several times, until you see no peak at the sensitivity level you’re interested in.
4. Then inject your sample solution and record the peak area.
5. Inject again the sample solvent and record the peak area – if any.
6. Divide the peak area of the solvent/blank peak by the peak area of the sample peak and multiply by 100.
7. The resultant value is percentage carry-over that is relevant to this particular method of analysis.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 1888
- Joined: Fri Aug 08, 2008 11:54 am
danko's right. The obvious approach is to include a blank runs after your highest calibration points, and look to see whether the peaks in these blanks are significant compared to the lowest level samples with which you expect to work. If they aren't, you are probably fine.
-
- Posts: 71
- Joined: Thu Jun 11, 2009 9:42 am
Completely agree. This is exactly the same procedure we apply in our validation protocols.danko's right. The obvious approach is to include a blank runs after your highest calibration points, and look to see whether the peaks in these blanks are significant compared to the lowest level samples with which you expect to work. If they aren't, you are probably fine.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy
Developability Dept.
NiKem Research S.r.l.
Italy
-
- Posts: 362
- Joined: Tue Aug 31, 2004 6:54 am
You might inject the solvent a second time. This actually would show you any contamination in the solvent.
A contamination in the solvent could otherwise idicate a too high carry-over.
A contamination in the solvent could otherwise idicate a too high carry-over.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
-
- Posts: 51
- Joined: Mon Jan 26, 2009 5:52 pm
I AGREE WITH IMH.....INJECT A BLANK AFTER YOUR HIGHEST CONCENTRATION ....
6 posts
Page 1 of 1
Who is online
In total there are 62 users online :: 1 registered, 0 hidden and 61 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 61 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 61 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science