Naming of Peaks in Chemstation

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Hi all,

I just wondered if anyone knows of a way to 'forcefully' name a peak in chemstation? I know that there are high / low RT columns in the calibration table but this isn't working as our peaks are drifting around a lot during the course of the run. As a result chemstation is naming some peaks wrongly causing problems with subsequent calculations in Chemstore. Anyone got any ideas?

Cheers,

JKB84
our peaks are drifting around a lot during the course of the run.
I think this is the problem you need to fix. If you are running gradient, allow more equilibration volume/time. If running buffer, is there enough buffer capacity? Small leak? Is your pressure constant in isocratic? Is there a column oven controlling temperature? Is the mobile phase degassed? Are you using a pre-mixed mobile phase and one component is evaporating preferentially, changing its composition over time? Stuff like that.
chemstation is naming some peaks wrongly causing problems with subsequent calculations in Chemstore.
I don't believe ChemStore does any calculations, just electronic archive of data for data integrity; fix the chromatography.

Can you clarify what you mean by drifting around?. If it is variable RT during a runs of similar products, then yes, as CPG says - fix the chromatography.

If you are running sequence of samples of fractions from a chromatography column, it's not unusual for some peaks to appear and grow, whilst neighbouring peaks shrink and disappear during the sequence. That can screw up peak identities.

In that case, the peak RT windows have to be closely specified, and you may have to select one, or more, retention time reference peak, or add internal standard.

As far as I'm aware, unless later versions have changed, Chemstore performed advanced reprocessing calculations, as well as statistics and trends.

Bruce Hamilton
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 2 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry